Duplicate LLOQ and ULOQ [Bioanalytics]

posted by Helmut Homepage – Vienna, Austria, 2012-01-20 14:09 (5256 d 20:05 ago) – Posting: # 7980
Views: 6,156

Dear Ohlbe!

❝ What most people do is that they only use one of the two replicates of the LLOQ and ULOQ samples. Usually the first replicate is used,…


Really most people? That’s strange to me. Why the first and not the second? Why not both?

❝ Some people take whichever value gives the highest R2, but I don't like that (I don't consider R2 to be a relevant parameter).


Right. R² is meaningless in calibration (if you prefer the statistical term: inverse regression). A nice reference (based on backcalculated concentrations):

Almeida AM, Castel-Branco MM, and AC Falcão
Linear regression for calibration lines revisited: weighting schemes for bioanalytical methods
Journal of Chromatography B 774, 215–22 (2002)