Selectivity [Bioanalytics]

posted by Ohlbe – France, 2018-11-13 14:56 (1984 d 20:21 ago) – Posting: # 19596
Views: 2,781

Dear Avinash,

❝ Which means? Should we run Selectivity samples with calibrators?


Not necessarily a full calibration curve, but at least samples spiked at the LLOQ, so that you can compare the signal in the blanks to the signal at the LLOQ.

Regards
Ohlbe