Calibration and QC point with same concentration [Bioanalytics]

posted by Relaxation  – Germany, 2017-02-24 11:29 (2589 d 01:30 ago) – Posting: # 17106
Views: 9,796

Hi.

❝ According to my boss, its "common sense" that the conc. shouldn't be same.


Hm, I took a different message home from this thread.

❝ Total CC points was eight so I think we can reprocess MV data with the problematic point removed and the curve will still have enough CC levels.


Actually, I even had a short discussion with a few (OK, 2, but both are no newbies) of my co-workers and one of them had an argument pro different concentrations (basically "I see this all the time") in case of clinical (!) laboratories.
But both (and me) would discourage omission of valid data from the calibration without proper justification (read: would start a discussion on re-calculating/re-validating including all CCs if we saw that in a bioanalytical/validation report).
Just my thoughts.


Best regards,

Steven.

Complete thread:

UA Flag
Activity
 Admin contact
22,957 posts in 4,819 threads, 1,639 registered users;
74 visitors (0 registered, 74 guests [including 7 identified bots]).
Forum time: 13:00 CET (Europe/Vienna)

Nothing shows a lack of mathematical education more
than an overly precise calculation.    Carl Friedrich Gauß

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5