lightsensitive drugs (suggestions) [Bioanalytics]

posted by Helmut Homepage – Vienna, Austria, 2007-05-29 14:47 (6615 d 22:42 ago) – Posting: # 754
Views: 6,101

Hi drgopal!

Let’s consider 2 phases were degradation may occur: clinics and analytics.

Clinics
Generally lightsensitive drugs are well protected during sampling, because:
  1. Glas vials (vacutainers) are almost perfect shields for UV-radiation.
  2. The entry-depth of light into whole blood is in the range of a few millimeters only.
  3. If your compound has a similar absorption wavelength as compared to albumin, such a compound may be well protected after centrifugation in plasma / plastic tubes.
In my experience working in the clinical stage under light protection (e.g., sodium vapour lamps) leads to a lot of difficulties in venipuncture, sampling errors, etc.
Plasma vials (3) may be wrapped in aluminium foil or dip-coated.

If you are over-cautious, you may try to validate 1 & 2, but:
be prepared to difficulities anyway, because you would have to spike whole blood in vacutainers. IMHO, you end up with very variable results; therefore more than the usual 6 replicates must be considered.

Analytics
  1. Analytes in stock solutions and sample extracts are much more susceptible to light-degradation than plasma samples.
  2. You should validate all sample preparation steps under varying light conditions (day light through [closed!] windows, fluorescent light, dimmed light, sodium vapour light), and different light protection measures (glass vials, brown glass vials, PP vials, etc).
  3. Dont’t forget to close the lid of the autosampler. ;-)
As in clinics, the dimmer the light the more sample preparation errors are likely to occur.
You should find an optimum between sufficient light protection and practicability.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes

Complete thread:

UA Flag
Activity
 Admin contact
23,428 posts in 4,929 threads, 1,686 registered users;
63 visitors (0 registered, 63 guests [including 21 identified bots]).
Forum time: 13:29 CEST (Europe/Vienna)

If I’d observed all the rules,
I’d never have got anywhere.    Marilyn Monroe

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5