Educate the IEC and regulators [Power / Sample Size]
Hi Yura,
As I wrote above the IEC and the authority should judge this before the study is done. I agree that in many cases the statistical knowledge of IECs is limited. However, once the protocol was approved by both, I don’t see a reason to talk about “forced BE” any more.
BTW, I don’t see a problem if a study is designed for 90% power (80% is not carved in stone). Let’s assume a dropout rate of 15% and we will already end up with 46 subjects:
Considering your example and assuming that the GMR and CV turn out exactly as assumed, no dropouts (n=50): The 90% CI will be 87.47–103.18%. Fine with me. Not even a significant difference (100% included). If the dropout-rate is as expected (n=38) the 90% CI will be 86.36–104.51%. If the assessor is not happy with that, he should have a chat with his colleague who approved the protocol and enlighten him about potential “over-powering” in study planing.
According to all guidelines (CI within the acceptance range) I can’t imagine a justification to reject the study. If the study is not accepted only due to the high sample size in the EEA the applicant might go for a referral (with extremely high chances of success) and in the USA the FDA will be sued right away.
❝ You are considering the power after the study at n = 28 (which were calculated before the study: GMR=0.95, CV=0.25, β=0.80). The question is whether it is possible to carry out a study at n = 50 and will this be forced bioequivalence?
As I wrote above the IEC and the authority should judge this before the study is done. I agree that in many cases the statistical knowledge of IECs is limited. However, once the protocol was approved by both, I don’t see a reason to talk about “forced BE” any more.
BTW, I don’t see a problem if a study is designed for 90% power (80% is not carved in stone). Let’s assume a dropout rate of 15% and we will already end up with 46 subjects:
library(PowerTOST)
CV <- 0.25 # CV-intra
theta0 <- 0.95 # T/R-ratio
target <- 0.90 # desired (target) power
dor <- 15 # expected dropout rate in percent
n <- sampleN.TOST(CV=CV, theta0=theta0, targetpower=target,
print=FALSE)[["Sample size"]]
ceiling(n/(1-dor/100)/2)*2 # round up to next even
Considering your example and assuming that the GMR and CV turn out exactly as assumed, no dropouts (n=50): The 90% CI will be 87.47–103.18%. Fine with me. Not even a significant difference (100% included). If the dropout-rate is as expected (n=38) the 90% CI will be 86.36–104.51%. If the assessor is not happy with that, he should have a chat with his colleague who approved the protocol and enlighten him about potential “over-powering” in study planing.
According to all guidelines (CI within the acceptance range) I can’t imagine a justification to reject the study. If the study is not accepted only due to the high sample size in the EEA the applicant might go for a referral (with extremely high chances of success) and in the USA the FDA will be sued right away.

—
Dif-tor heh smusma 🖖🏼 Довге життя Україна!![[image]](https://static.bebac.at/pics/Blue_and_yellow_ribbon_UA.png)
Helmut Schütz
![[image]](https://static.bebac.at/img/CC by.png)
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Dif-tor heh smusma 🖖🏼 Довге життя Україна!
![[image]](https://static.bebac.at/pics/Blue_and_yellow_ribbon_UA.png)
Helmut Schütz
![[image]](https://static.bebac.at/img/CC by.png)
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Complete thread:
- Power is getting high kms.srinivas 2017-12-26 07:39
- Power is getting high? d_labes 2017-12-26 12:08
- Stop estimating post hoc power! Helmut 2017-12-26 12:22
- Stop estimating post hoc power! kms.srinivas 2017-12-26 13:00
- Stop estimating post hoc power! ElMaestro 2017-12-26 14:11
- Simulations Helmut 2017-12-26 15:51
- Simulations BE-proff 2017-12-27 06:53
- Simulations ElMaestro 2017-12-27 07:25
- Simulations Yura 2017-12-27 08:23
- Simulations kms.srinivas 2017-12-27 09:11
- Simulations Yura 2017-12-27 10:07
- “Forced BE” 101 Helmut 2017-12-27 12:23
- “Forced BE” 101 kms.srinivas 2017-12-27 12:41
- Would you be so kind answering our questions? Helmut 2017-12-27 13:02
- Would you be so kind answering our questions? kms.srinivas 2017-12-28 05:53
- Yes, but why? Helmut 2017-12-28 11:47
- Yes, but why? DavidManteigas 2017-12-28 16:59
- Optimists and pessimists Helmut 2017-12-28 17:33
- "normal" GMR setting d_labes 2017-12-28 18:57
- Example for discussion mittyri 2017-12-28 22:06
- Example for discussion Helmut 2017-12-28 22:33
- I prefer to play it safe Helmut 2017-12-28 22:10
- Example for discussion mittyri 2017-12-28 22:06
- "normal" GMR setting d_labes 2017-12-28 18:57
- Full ACK d_labes 2017-12-28 17:41
- About GMR 1.10 kms.srinivas 2017-12-29 13:20
- Better 0.95 or 0.90 Helmut 2017-12-29 16:18
- Optimists and pessimists Helmut 2017-12-28 17:33
- Yes, but why? Yura 2017-12-29 13:46
- Buffon's needle Astea 2018-01-20 23:55
- Buffon's needle Oleg777 2018-10-09 22:48
- 0.95 or 1.05 Helmut 2018-10-10 13:41
- Buffon's needle Helmut 2018-10-10 12:46
- Buffon's needle Astea 2018-10-11 23:14
- School maths Helmut 2018-10-12 01:10
- School russian Astea 2018-10-12 12:41
- Offtop: Umschrift der westlichen Eigennamen auf Russisch mittyri 2018-10-12 23:25
- School maths Helmut 2018-10-12 01:10
- Buffon's needle Astea 2018-10-11 23:14
- Buffon's needle Oleg777 2018-10-09 22:48
- Yes, but why? DavidManteigas 2017-12-28 16:59
- Yes, but why? Helmut 2017-12-28 11:47
- EEU? mittyri 2017-12-28 21:52
- EEU? Yura 2017-12-29 13:41
- EEU - pharmacokinetic equation??? mittyri 2017-12-29 14:11
- EEU? Beholder 2018-01-16 15:10
- EEU? Yura 2017-12-29 13:41
- Would you be so kind answering our questions? kms.srinivas 2017-12-28 05:53
- Would you be so kind answering our questions? Helmut 2017-12-27 13:02
- “Forced BE” 101 kms.srinivas 2017-12-27 12:41
- Simulations kms.srinivas 2017-12-27 09:11
- Simulations xtianbadillo 2018-01-18 22:22
- Simulations BE-proff 2017-12-27 06:53
- Stop estimating post hoc power! kms.srinivas 2017-12-26 13:00
- Numbers don't lie ElMaestro 2017-12-28 20:13